[FDS] A couple of observations
Tom Laue
Tom.Laue at unh.edu
Sun Oct 29 14:47:07 CST 2006
Hi all-
I've had a chance to work with the FDS system a bit this fall, and have
a couple of observations.
1. Free fluorescein-type dyes seems to sediment at ~0.1 to 0.2. We
see this in nearly every sample we analyze.
1. To determine the amount of free dye using c(s) analysis, set
the F-ratio to 0 (see attachment). Otherwise, the dye peak
will be smeared into the 0 s hook.
2. I've attached a snapshot from a c(s) analysis of a
fluorescein-labeled protein. The sample had been passed over
a desalting column prior to analysis. I'm not sure whether
the dye has hydrolyzed from the protein, or if it is carried
over the desalting column by non-covalent interactions.
Anyone have an idea about this? There are a couple of
comments on the web (e.g.
http://www.microbiology.emory.edu/altman/jdaWebSite_v3/p_commOnRoederer.shtml#bufferX)
indicating there may be incomplete free dye removal on
desalting columns. I do not know if this is a problem generally.
2. The object lens focus must be set near the right level before
aligning the dichroic mirror for optimal signal.
1. A symptom that the dichroic mirror is out of alignment is
that a Focus scan (reached from Fluorescence/Fluorescence
focusing...) produces a "double humped" intensity profile.
2. The correct focus position is usually at the minimum between
the two humps
3. If you choose a lens position corresponding to one of the
two "humps," alignment of the mirror for optimal intensity
will not get rid of the double humps
4. Instead, choose a lens position between the two humps, then
align the dichroic mirror to optimize for intensity. You
will wind up with a stronger signal and good radial resolution
Best wishes,
Tom
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Rudman Hall 379
46 College Road
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
www.camis.unh.edu
www.bitc.unh.edu
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://biochem.uthscsa.edu/pipermail/fds/attachments/20061029/4654734e/attachment-0001.html
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Free dye.pdf
Type: application/pdf
Size: 146184 bytes
Desc: not available
Url : http://biochem.uthscsa.edu/pipermail/fds/attachments/20061029/4654734e/attachment-0001.pdf
More information about the FDS
mailing list